Reference Guide

Hemodynamic Monitoring

Swan-Ganz catheter, arterial lines, CVP, cardiac output, and shock patterns

Reference Values

ParameterNormal Range
Right Atrial Pressure (RAP/CVP)−1 to +8 mmHg
Right Ventricular Pressure15–28 / 0–8 mmHg
Pulmonary Artery Pressure15–28 / 5–16 mmHg; mean 10–22 mmHg
Pulmonary Capillary Wedge Pressure (PCWP)6–15 mmHg (mean)
Cardiac Output (CO)4.8–7.3 L/min
Cardiac Index (CI)2.8–4.2 L/min/m²
Systemic Vascular Resistance (SVR)700–1600 dyn·s·cm⁻⁵
Pulmonary Vascular Resistance (PVR)0.25–2 Wood units (20–130 dyn·s·cm⁻⁵)
Mixed Venous O₂ Saturation (SvO₂)65–75%
Cardiac Power Output (CPO)>0.6 W (low predicts MCS need)
PA Pulsatility Index (PAPi)>1.0 (low suggests RV failure)

Key Note

All pressures should be read at end-expiration, when intrathoracic pressure most closely approximates atmospheric pressure, regardless of ventilation mode.

The PAC provides direct measurement of right-sided pressures, PA pressures, and an estimate of left atrial pressure via the wedge, along with cardiac output by thermodilution. Despite debate (ESCAPE trial found no mortality benefit in routine use), contemporary reappraisal supports invasive hemodynamic profiling in cardiogenic shock and patients on mechanical circulatory support (MCS).

Clinical Use Pearls

  • CPO = MAP × CO / 451; CPO <0.6 W predicts mortality and need for MCS escalation in cardiogenic shock
  • PAPi = (PA systolic − PA diastolic) / RAP; PAPi <1.0 (especially <0.9 in acute MI) predicts RV failure and need for right-sided MCS
  • Transpulmonary gradient (mean PAP − PCWP) >12 mmHg suggests pulmonary vascular disease
  • Cardiac index = CO / BSA; thermodilution curves should be reproducible within ~10% across three injections
  • Continuous CO (CCO) averages over 4–12 min; STAT CCO updates every 30–60 seconds
  • Limitations: thermodilution unreliable with severe TR, intracardiac shunts, or arrhythmias

Setup & Zeroing

  • Transducer leveling: Phlebostatic axis (4th intercostal space, midaxillary line)
  • Re-zero whenever the system is opened to air or readings are questioned
  • Tubing: Short (<122 cm), stiff, non-compliant; remove all air bubbles
  • Pressure bag: Maintain ≥300 mmHg

Square Wave (Fast-Flush) Test

ResponseOscillationsCauseClinical Effect
Optimally Damped1–2Properly set up systemAccurate readings
Underdamped>2Air bubbles, long/compliant tubing, catheter whipFalsely HIGH SBP, falsely LOW DBP
Overdamped0–1 (sluggish)Clot, kink, large air bubble, loose connectionFalsely LOW SBP, falsely HIGH DBP; loss of dicrotic notch

Troubleshooting Sequence

Check stopcocks → Check pressure bag (≥300 mmHg) → Flush line and tighten connections → Re-zero → Reposition catheter (especially radial site) → Consider replacement if persistently damped.

Normal Waveform Components

  • a wave – Atrial contraction; corresponds to RV end-diastole. Best approximates true RV filling pressure
  • c wave – Tricuspid valve closure/bulging into RA at onset of ventricular systole
  • x descent – Atrial relaxation; blunting suggests RV dysfunction
  • v wave – Passive atrial filling against a closed tricuspid valve
  • y descent – Atrial emptying after tricuspid opening

Abnormal Findings

AbnormalityInterpretation
Loss of a wavesAtrial fibrillation
Flutter a wavesAtrial flutter
Cannon a wavesAV dissociation, junctional/ventricular rhythm, complete heart block
Giant v wavesTricuspid regurgitation; acute MR reflected at PCWP
Blunted x descentRV dysfunction
M or W pattern (steep x and y)Constrictive pericarditis
Blunted y descentCardiac tamponade
Arterial-appearing waveformCatheter migration into RV — reposition immediately

Methods Comparison

MethodPrincipleStrengthsLimitations
PAC Thermodilution (bolus)Cold saline RA → PA thermistorGold standard; well-validatedOperator-dependent; inaccurate with TR, shunts
Continuous CO (PAC)Thermal filament algorithmTrending without injections4–12 min lag; STAT 30–60 s
Transpulmonary (PiCCO)Central venous → arterial thermistorProvides EVLW, GEDVRequires femoral arterial line
Pulse Contour (calibrated)Arterial waveform + calibrationContinuous; less invasiveDrifts; needs recalibration
Pulse Contour (uncalibrated)Waveform algorithm onlyMinimally invasivePoor in low EF, high-dose vasopressors
Fick (calculated)VO₂ / (CaO₂ − CvO₂) × 10Useful in shunts, severe TRRequires assumed VO₂ or metabolic cart
Echo (LVOT VTI)LVOT area × VTI × HRNoninvasiveOperator-dependent

Hemodynamic Profiles

Shock StateCISVRCVPPCWPSvO₂
Hypovolemic
Cardiogenic (LV)
Cardiogenic (RV)↑↑↓ or N
Tamponade↑ (equalized)↑ (equalized)
Massive PE
Distributive/Septic↑ or N↓↓↓ or N↓ or N↑ or N
Mixed (cardiogenic + vasoplegic)

RV Failure Pattern

  • CVP/RAP >15 mmHg, CVP/PCWP ratio >0.8
  • PAPi <1.0, low PA pulse pressure, blunted x descent
  • Dilated/hypokinetic RV on echo, septal shift
  • Avoid volume loading once CVP is high — worsens RV dilation and ventricular interdependence
  • Optimize rhythm (sinus, AV synchrony), reduce RV afterload (treat hypoxia/hypercapnia/acidosis)
  • Add inotropy (milrinone, dobutamine, epinephrine), inhaled pulmonary vasodilators
  • Consider RV MCS (RVAD, Impella RP, VA-ECMO) if refractory

Clinical Pearls

  • Read all hemodynamic pressures at end-expiration
  • A normal CVP does not exclude RV failure — look at PAPi, x descent, RV size, CVP/PCWP ratio
  • Cannon a waves on CVP in a bradycardic patient → AV dissociation — check pacer and rhythm
  • New giant v wave on PCWP → acute mitral regurgitation (e.g., papillary muscle rupture)
  • Falling SvO₂ (<60%) is an early sign of inadequate O₂ delivery — evaluate CO, Hgb, SaO₂, VO₂
  • Sudden cessation of chest tube output + rising filling pressures + narrowing pulse pressure = tamponade until proven otherwise
  • Hypotension after intubation → loss of sympathetic tone + decreased preload from PPV — treat with fluid, vasopressor, minimize PEEP
  • In RV failure, volume is the enemy — favor diuresis, inotropy, and afterload reduction
  • Always confirm pacer capture and sensing at start of every shift and before transport